Skip to main content
MedVellum
MCQsExamsAtlas
DashboardPricing
MBBS / Core medicine✳Dermatology✳ICU Fellowship (CICM)✳Anaesthesia✳Emergency Medicine✳Psychiatry Fellowship✳Paediatrics Fellowship✳Physician Medicine✳MCQs✳SAQs✳Vivas✳OSCE✳Evidence-first✳MBBS / Core medicine✳Dermatology✳ICU Fellowship (CICM)✳Anaesthesia✳Emergency Medicine✳Psychiatry Fellowship✳Paediatrics Fellowship✳Physician Medicine✳MCQs✳SAQs✳Vivas✳OSCE✳Evidence-first✳

MedVellum.

The folio

Exam-exhaustive medical education across every specialty — evidence-graded topics, engraved plates, and practice in every written and oral format. Educational content only — not medical advice.

llms.txt · psychiatry LLM catalog · sitemap

Atlas

  • Specialty atlas
  • MBBS / Core medicine
  • Dermatology
  • ICU Fellowship (CICM)
  • Anaesthesia
  • Emergency Medicine
  • Psychiatry Fellowship
  • Paediatrics Fellowship
  • Physician Medicine

Study & account

  • MCQ practice
  • Practice alias
  • Exam tools
  • Dashboard
  • Pricing
  • Sign in

© 2026 MedVellum. For education only — not a substitute for clinical judgement.

Folio edition · Set in Instrument Serif & Archivo

Anaes TopicsPerioperative medicine

Anaes · Perioperative medicine

Perioperative management of pacemakers and ICDs

Also known as CIED perioperative management · Pacemaker magnet mode · ICD disable for surgery · DDD pacing anaesthesia · Electromagnetic interference pacemaker

Exam-exhaustive CIED perioperative care: NBG modes at high level, pacing dependence, magnet behaviour (pacemaker vs ICD), diathermy EMI mitigation, reprogramming indications, ICD therapy suspension, external defibrillation readiness, and post-op interrogation for ANZCA Final and equivalents.

high3 referencesUpdated 10 July 2026
On this page & tools

Your progress

Saved locally on this device.

Target exams

ANZCAFRCAABAEDAICFCAI

Red flags

Magnet on a pacemaker → asynchronous pacing; magnet on an ICD → suspends tachy therapy but does NOT make pacing asynchronous — know the difference.Monopolar diathermy can inhibit pacing or trigger inappropriate ICD shocks — prefer bipolar, short bursts, return pad path away from generator.Pacing-dependent + EMI without a plan = asystole. Have magnet, external pads, and cardiology contact before knife-to-skin.Suspending ICD therapy without external defibrillator capability is incomplete preparation.Post-op: re-enable ICD therapies and re-interrogate if settings changed or EMI exposure was significant.

Your progress

Saved locally on this device.

Target exams

ANZCAFRCAABAEDAICFCAI

Red flags

Magnet on a pacemaker → asynchronous pacing; magnet on an ICD → suspends tachy therapy but does NOT make pacing asynchronous — know the difference.Monopolar diathermy can inhibit pacing or trigger inappropriate ICD shocks — prefer bipolar, short bursts, return pad path away from generator.Pacing-dependent + EMI without a plan = asystole. Have magnet, external pads, and cardiology contact before knife-to-skin.Suspending ICD therapy without external defibrillator capability is incomplete preparation.Post-op: re-enable ICD therapies and re-interrogate if settings changed or EMI exposure was significant.

Key answer

Identify the device and pacing dependence, plan EMI (bipolar/short monopolar bursts, return pad path), decide magnet vs formal reprogramming, suspend ICD tachy therapies with external pads ready, and re-enable/re-interrogate afterwards.
[1]
Educational illustration of cardiac implantable electronic device perioperative management
FigureCIED perioperative spine: know the device, protect from EMI, magnet vs reprogram, defibrillator readiness

Why this is examined / the one-line answer

Almost every general list has a patient with a pacemaker or ICD. Examiners test whether you can decode the problem, not recite cardiology textbooks. The viva turns on four discriminators: what does the device do, is the patient pacing-dependent, what will diathermy do to it, and magnet versus reprogramming versus ICD therapy off.[1][2]

One-liner: I confirm device type and indication, assess pacing dependence, minimise EMI, use magnet or reprogramming as indicated, disable ICD shock therapies with external defibrillation ready, and restore/check the device after surgery. [1]

ASA 2020 Practice Advisory and HRS/ASA consensus provide the named frameworks examiners expect.[1][2]

Preoperative assessment and risk stratification

What you must know before induction

ItemWhy it changes the plan
Device typePacemaker only vs ICD (± pacing) vs CRT-P/CRT-D vs leadless/S-ICD
IndicationSinus node disease, AV block, primary/secondary prevention SCD, heart failure
Manufacturer / modelMagnet response varies; interrogation software differs
Implant date and last checkBattery, lead thresholds, recent shocks
Pacing dependenceNo reliable intrinsic rhythm → EMI inhibition is catastrophic
Rate-response sensorsEMI/vibration may drive inappropriate high-rate pacing
Special featuresMinute ventilation sensors, antitachycardia pacing, subcutaneous ICD

Pacing dependence (exam definition): absence of a reliable haemodynamically adequate intrinsic escape rhythm. Clues: underlying complete heart block, prior asystole when pacing suspended at clinic, 100% paced on ECG/device report, symptoms when rate falls. When unsure, treat as dependent until proven otherwise.[3]

Preoperative interrogation — when mandatory / strongly preferred

  • Elective major surgery with significant EMI risk (monopolar above umbilicus, near generator)
  • ICD patients when therapy will be suspended
  • Unknown device function, recent shocks, battery alerts, or long interval since check
  • Pacing-dependent patients
  • Emergency: do not delay life-saving surgery for formal programming if magnet + pads + monitoring can temporise — document and arrange earliest interrogation.[1][2]

Ask the patient for their device card. If unavailable, CXR may show generator/leads; ECG may show pacing spikes. [1]

Investigations and optimisation

Recent electrolytes (especially K+ — hypo/hyperkalaemia alter capture thresholds), ECG, consider echo if heart failure/CRT, continue usual cardiac meds unless specific reason to hold. Document anticoagulation if recent lead revision (rare acute issue). [1]

Applied physiology / device logic (high-yield, not encyclopaedic)

NBG code — the letters you must decode

Pacemaker modes are classically three (or more) letters: [1]

  1. Chamber paced: O = none, A = atrium, V = ventricle, D = dual (A+V)
  2. Chamber sensed: O / A / V / D
  3. Response to sensing: O = none, T = triggered, I = inhibited, D = dual (triggered + inhibited)
  4. (Often) Rate modulation: R = rate-responsive
  5. (Sometimes) Multisite pacing [1]

Exam workhorses: [1]

ModePlain-EnglishPerioperative note
DDDDual chamber pacing and sensing; tracks atrium, paces ventricle if neededCommon modern mode; EMI may inhibit or track noise
VVIVentricular demand pacingSimple; EMI can inhibit → asystole if dependent
AAIAtrial demandIntact AV conduction assumed
DOO / VOOAsynchronous dual/ventricular pacingMagnet or programmed EMI-safe mode — no sensing of noise
DDDR / VVIRRate-responsiveSensors may misread EMI/shivering as exercise

You do not need every obscure code. You need: sensing can be fooled by EMI; asynchronous modes ignore EMI for pacing decisions; ICDs add a shock layer on top. [1]

EMI mechanisms (theatre focus)

Monopolar diathermy is the classic source. Noise may be sensed as: [1]

  • Intrinsic QRS → pacing inhibition (dangerous if dependent)
  • Tachyarrhythmia → ICD charge/shock (inappropriate)
  • Rate-response drive → inappropriate tachycardia
  • Rare: reset/reprogram, or power-on reset to backup mode [1]

Bipolar diathermy has a short current path and is preferred when surgically acceptable. Ultrasonic scalpels have lower EMI risk than monopolar RF but are not zero-risk.[2][3]

Classification of CIED types and EMI interaction modes
FigureDevice class and EMI effect: inhibition, inappropriate shock, rate-response overdrive, rare reset

Anaesthetic goals

  1. Maintain capture and cardiac output — never leave a dependent patient without a pacing plan.
  2. Minimise EMI or make the device ignore it (asynchronous pacing / ICD therapy off).
  3. Keep external defibrillation/pacing immediately available when ICD therapies suspended.
  4. Avoid extremes of K+, acid–base, and ischaemia that raise capture threshold.
  5. Restore correct device function before unsupervised ward care. [1]

Technique options and decision matrix

Magnet vs reprogramming

Magnet teaching mantra: Pacemaker magnet → asynchronous pacing. ICD magnet → shock therapies off, pacing usually unchanged. Always state you will confirm magnet behaviour against manufacturer/device report because exceptions exist.[1][3]

When to reprogram (high-yield list)

  • Pacing-dependent patient + monopolar EMI likely near generator/leads
  • ICD + surgery with sustained EMI risk (suspend detection/therapies)
  • Rate-response or minute-ventilation sensors that may malfunction with EMI/ventilation changes
  • Unreliable magnet access (prone, generator site in field, sterile exclusion)
  • Complex CRT or recent device issues [1]

Diathermy practical rules (say them cleanly)

  1. Prefer bipolar or ultrasonic where possible.
  2. If monopolar essential: place return pad so current path does not cross generator/heart (e.g. ipsilateral thigh for arm surgery away from generator path).
  3. Keep active electrode away from generator (classic teaching distance often quoted around 15 cm — state principle, not magic number alone).
  4. Short intermittent bursts (commonly taught <5 s), lowest effective power.
  5. Full monitoring; watch for missing paced beats or VT/VF alarms.
  6. Magnet immediately available; programmer if planned.[2][3]

ICD disable for surgery — the exam algorithm

  1. Confirm ICD (transvenous or S-ICD).
  2. Risk of EMI → suspend tachy therapies (reprogram preferred for planned major EMI; magnet if urgent/short).
  3. Apply external multifunction pads in axis that avoids generator when possible; connect to defibrillator.
  4. Continuous ECG (note: device magnet may alter pacing display).
  5. After surgery: re-enable therapies and document; interrogate if any concern.
  6. Until therapies restored, patient needs monitored environment with defibrillator access.[1][2]

ICD off without pads is incomplete

Suspending ICD therapies protects against inappropriate shocks from EMI but leaves the patient unprotected from true VT/VF — external defibrillation readiness is mandatory until therapies are restored.

[1]

Monitoring and equipment

  • Standard ASA monitoring + attention to pulse oximeter / arterial waveform as mechanical pulse confirmation if ECG is noisy
  • External pads on before incision when ICD therapies off or high EMI risk
  • Magnet labelled and in theatre
  • CIED team / cardiology contact number
  • Emergency drugs for arrhythmia and support (adrenaline, amiodarone per ALS, isoprenaline rarely for brady if capture lost — but external pacing/defib first)
  • Avoid placing central lines / subclavian wires blindly ipsilateral to fresh leads without indication awareness (lead injury risk) [1]

Intraoperative management

Perioperative CIED management pathway
FigureManagement pathway: assess dependence → EMI plan → magnet/reprogram/ICD off → pads ready → restore and recheck

Induction: standard agents; avoid profound bradycardia/hypotension in low EF CRT patients. Succinylcholine fasciculations rarely discussed as EMI-like noise — usually not a practical issue with modern devices but beware in dependent patients if ECG noise appears. [1]

Maintenance: communicate with surgeon on diathermy timing. If inhibition occurs: stop diathermy, apply magnet (pacemaker), support circulation, call for help/programmer. [1]

Positioning: prone cases — magnet may not stay; prefer reprogramming. [1]

Lithotripsy / RF ablation / ECT / MRI: special protocols — generally disable ICD therapies, protect generator from shock wave focus, MRI only if conditional device + protocol (cross-link MRI topic). [1]

Crisis pivots

Asystole / severe bradycardia during diathermy

Stop EMI → 100% oxygen → magnet for pacemaker → external pacing/pads → CPR if needed → emergency reprogramming → correct reversible causes (K+, ischaemia, drugs). [1]

Inappropriate ICD shock under anaesthesia

Stop EMI → magnet to suspend further therapies if not already → treat any resultant arrhythmia → interrogate → continue only with therapies managed and pads on. [1]

Failure to capture

Check connections/ECG lead-off vs true failure; optimise electrolytes; increase external pacing output; urgent cardiology — possible lead issue or threshold rise. [1]

Power-on reset / unexpected mode

Assume backup mode (often VVI-like at fixed rate); arrange urgent interrogation; do not discharge to unmonitored ward. [1]

Postoperative / PACU plan

  • Re-enable ICD therapies before leaving monitored care whenever suspended
  • Re-interrogate after reprogramming, major EMI, shocks, or haemodynamic events
  • Document mode, therapies status, and next cardiology follow-up
  • Ward: continuous monitoring until device function confirmed if any doubt
  • Analgesia standard; no special CIED interaction with most analgesics [1]

Special populations and devices

CRT-D: loss of LV pacing may drop output — haemodynamic vigilance. [1]

Leadless pacemaker: still EMI-aware; magnet/programmer specifics differ — check manufacturer pathway. [1]

S-ICD: no transvenous pacing (usually); magnet/therapy suspend still relevant for shocks; pads placement careful. [1]

Paediatric CIEDs: smaller generators, growth-related leads — same EMI principles, lower threshold for specialist involvement. [1]

Emergency surgery at night: magnet + pads + bipolar preference + earliest daytime interrogation is acceptable temporising strategy when programmer unavailable.[1]

SAQ answer scaffold

A 72-year-old with a dual-chamber pacemaker for complete heart block presents for laparoscopic hemicolectomy. How will you manage the CIED perioperatively? [1]

  1. Assess (3): device card/interrogation, pacing dependence (likely high), battery/leads, ECG.
  2. EMI plan (3): prefer bipolar; if monopolar, return pad path, short bursts, distance from generator.
  3. Device strategy (4): asynchronous pacing via magnet or reprogramming; rate-response off if programmed; continuous monitoring + external pads available.
  4. Post-op (2): restore original settings; re-interrogate; monitored recovery until safe.[1][2]

Viva stem bank and model phrases

Stem 1: “What does DDD mean?”
Model: “Dual chamber paced, dual chamber sensed, dual response — it paces and senses both atrium and ventricle with inhibition and tracking as programmed.” [1]

Stem 2: “Magnet on ICD?”
Model: “In most ICDs a magnet suspends tachyarrhythmia detection and therapies; it does not convert pacing to asynchronous mode the way a pacemaker magnet typically does.”[3]

Stem 3: “Patient is pacing-dependent — diathermy causes asystole.”
Model: “Stop diathermy, apply magnet for asynchronous pacing, external pace/defibrillate as needed, support ABC, arrange emergency programming.” [1]

Stem 4: “Surgery below the umbilicus — any issue?”
Model: “Lower EMI risk if current path remote, but monopolar can still affect devices; I still identify dependence and keep a magnet available.”[2]

Stem 5: “When must you reprogram?”
Model: “When magnet is impractical, when ICD therapies need reliable suspension for major EMI, when sensors are problematic, or when dependence plus high EMI risk demands a programmed asynchronous mode.” [1]

Stem 6: “Where do you put the diathermy plate?”
Model: “So the current path between active electrode and return pad does not cross the generator and heart.” [1]

Stem 7: “ICD therapies off — destination?”
Model: “Monitored bed with defibrillator until therapies re-enabled and device checked.” [1]

Common traps

  • Treating ICD magnet like pacemaker magnet (assuming asynchronous pacing)
  • Leaving ICD off on the ward
  • No external pads when therapies suspended
  • Long continuous monopolar bursts over the chest
  • Ignoring pacing dependence
  • Forgetting to restore settings after reprogramming
  • Assuming “below umbilicus = no risk” without thinking [1]

Red flag

A pacing-dependent patient exposed to monopolar EMI without asynchronous pacing protection can arrest from pacing inhibition — stop diathermy, magnet, external pace.
[1]

Clinical pearl

Say out loud at team brief: device type, dependence, magnet vs reprogram plan, ICD therapies on/off, and who re-enables the device. The brief is the safety intervention.
[1]

CIED theatre brief — PACE

[1]
ASA CIED 2020
Key advisory
Async pacing
Pacemaker magnet
Therapies suspend
ICD magnet
Bipolar
Preferred diathermy
Re-enable + pads until then
Post ICD-off

Examiner mental map

  1. Identify device and dependence.
  2. Predict EMI.
  3. Choose magnet vs reprogram.
  4. Protect ICD patient with therapy suspend + external defib.
  5. Mitigate diathermy.
  6. Restore and document.
    Hit those six without waffle and you pass the CIED viva. [1]

References

  1. [1]American Society of Anesthesiologists Practice Advisory for the Perioperative Management of Patients with Cardiac Implantable Electronic Devices: Pacemakers and Implantable Cardioverter-Defibrillators 2020: An Updated Report by the American Society of Anesthesiologists Task Force on Perioperative Management of Patients with Cardiac Implantable Electronic Devices: Erratum Anesthesiology, 2020.PMID 32032098
  2. [2]Crossley GH, Poole JE, Rozner MA, et al. The Heart Rhythm Society (HRS)/American Society of Anesthesiologists (ASA) Expert Consensus Statement on the perioperative management of patients with implantable defibrillators, pacemakers and arrhythmia monitors: facilities and patient management this document was developed as a joint project with the American Society of Anesthesiologists (ASA), and in collaboration with the American Heart Association (AHA), and the Society of Thoracic Surgeons (STS) Heart Rhythm, 2011.PMID 21722856
  3. [3]Stone ME, Salter B, Fischer A Perioperative management of patients with cardiac implantable electronic devices Br J Anaesth, 2011.PMID 22156267