Skip to main content
MedVellum
MCQsExamsAtlas
DashboardPricing
MBBS / Core medicine✳Dermatology✳ICU Fellowship (CICM)✳Anaesthesia✳Emergency Medicine✳Psychiatry Fellowship✳Paediatrics Fellowship✳Physician Medicine✳MCQs✳SAQs✳Vivas✳OSCE✳Evidence-first✳MBBS / Core medicine✳Dermatology✳ICU Fellowship (CICM)✳Anaesthesia✳Emergency Medicine✳Psychiatry Fellowship✳Paediatrics Fellowship✳Physician Medicine✳MCQs✳SAQs✳Vivas✳OSCE✳Evidence-first✳

MedVellum.

The folio

Exam-exhaustive medical education across every specialty — evidence-graded topics, engraved plates, and practice in every written and oral format. Educational content only — not medical advice.

llms.txt · psychiatry LLM catalog · sitemap

Atlas

  • Specialty atlas
  • MBBS / Core medicine
  • Dermatology
  • ICU Fellowship (CICM)
  • Anaesthesia
  • Emergency Medicine
  • Psychiatry Fellowship
  • Paediatrics Fellowship
  • Physician Medicine

Study & account

  • MCQ practice
  • Practice alias
  • Exam tools
  • Dashboard
  • Pricing
  • Sign in

© 2026 MedVellum. For education only — not a substitute for clinical judgement.

Folio edition · Set in Instrument Serif & Archivo

Anaes TopicsVaporisers

Anaes · Vaporisers

Vaporisers

Also known as Vaporiser · Plenum vaporiser · Variable-bypass · Draw-over vaporiser · Desflurane vaporiser · Temperature compensation

The vaporiser vaporises the liquid volatile anaesthetic agent and adds a precise concentration to the fresh gas flow. The framework rests on four exam-critical ideas: the variable-bypass plenum vaporiser (the standard for sevoflurane and isoflurane — a fraction of the gas is saturated in the chamber and diluted by the bypass); the temperature compensation (the bimetallic strip or the bellows compensates for the cooling of the vaporisation); the desflurane exception (its near-atmospheric saturated vapour pressure demands a heated, pressurised vaporiser); and the safety systems (the agent-specific filling, the interlock, the flow-resistance). Built on the AAGBI checking guidelines (2012), the anaesthetic machine assessment (Tiviraj 2016), and the closed-circuit review (Parthasarathy 2013).

medium3 referencesUpdated 10 July 2026
On this page & tools

Your progress

Saved locally on this device.

Target exams

ANZCAFRCAABAEDAICFCAIFCA_SA

Red flags

Desflurane boils at 22.8 degrees Celsius — its saturated vapour pressure (669 mmHg at 20 degrees) is near atmospheric. A standard variable-bypass vaporiser could let it boil and deliver a dangerously high concentration. Desflurane MUST use a heated, pressurised vaporiser.

Your progress

Saved locally on this device.

Target exams

ANZCAFRCAABAEDAICFCAIFCA_SA

Red flags

Desflurane boils at 22.8 degrees Celsius — its saturated vapour pressure (669 mmHg at 20 degrees) is near atmospheric. A standard variable-bypass vaporiser could let it boil and deliver a dangerously high concentration. Desflurane MUST use a heated, pressurised vaporiser.

Key answer

Variable-bypass plenum vaporiser: splitting ratio sends a fraction of FGF through a chamber to saturate at SVP, then remixes to dialled %. Temperature compensation (bimetallic strip/bellows) counters cooling of vaporisation. Agent-specific; keyed fillers; interlock one vaporiser on. Desflurane needs heated pressurised vaporiser because SVP near atmospheric and boils at 22.8 °C. Draw-over is low-resistance for remote/military use. [1]
[1]
Variable bypass vaporiser principle
FigureSplitting ratio + saturated chamber + temperature compensation = accurate volatile output.

Why this matters

If the vaporiser is wrong, the patient is too deep, too light, or receiving the wrong agent. Primary equipment exams live here [1][2].

Variable-bypass plenum principle

Fresh gas splits into: [1]

  1. Bypass flow — no agent
  2. Chamber flow — fully saturated with agent vapour at the prevailing SVP [2]

Streams reunite. Dial position changes the splitting ratio (fraction through chamber). Plenum vaporisers are pressurised by upstream flowmeters and present resistance — accurate across usual FGF rates when upright and filled correctly. [3]

Saturated vapour pressure table (20 °C order)

AgentSVP ~20 °CBoiling pointVaporiser type
Sevoflurane~157 mmHg (21 kPa)~59 °CVariable-bypass
Isoflurane~238 mmHg (32 kPa)~49 °CVariable-bypass
Halothane~243 mmHg~50 °CVariable-bypass
Desflurane~669 mmHg (89 kPa)22.8 °CHeated, pressurised special

Saturated chamber gas would contain (SVP/atmospheric) × 100% agent if undiluted — for desflurane that is ~90%, far above clinical 6–12%, and boiling risks uncontrolled output in a standard bypass design. [2]

Temperature compensation

Vaporisation is endothermic → liquid cools → SVP falls → output would fall. Compensation: [3]

  • Bimetallic strip in bypass narrows bypass when cold → more chamber flow
  • Aneroid bellows liquid expansion devices similarly adjust
  • Thermal mass / wicks maintain saturation [1]

Specify output stability within clinical temperature bands (teaching often ±15% of dialled across 15–35 °C). [2]

Safety features

FeaturePurpose
Agent-specific calibrationSVP and splitting differ by agent
Keyed fillers (colour/geometry)Prevent wrong liquid in wrong vaporiser
Interlock on back barOnly one vaporiser open
Anti-spill / level windowsAvoid overfill and tipping hazards
Downstream placementAfter flowmeters, before common gas outlet
Selectatec / interlocking mountingMechanical interlock compatibility

Desflurane vaporiser (Tec 6 type principles)

Plenum versus desflurane versus draw-over
FigureThree families: variable-bypass plenum, heated desflurane, and low-resistance draw-over.
FeatureWhy
Heated to ~39 °CRaise SVP / control boiling
Pressurised chamberControlled pure vapour injects into FGF
Electronics / alarmsPower failure ceases output safely
Not a simple variable-bypassBecause room-temp physics are hostile

Draw-over vaporisers

Low resistance, often not temperature/flow compensated to the same degree, used with self-inflating bag / bellows in field anaesthesia (e.g. Oxford Miniature Vaporiser concepts). Output varies more with FGF and temperature — operator skill matters. [2]

Altitude note

Variable-bypass vaporisers deliver roughly constant partial pressure output with altitude changes more than constant volume percent in some analyses — exams may ask whether partial pressure or volume % is preserved; clinical potency tracks partial pressure. Desflurane electronic units differ — know the principle that partial pressure drives anaesthesia. [3]

Output estimation (conceptual)

If chamber saturates at fraction F_sat = SVP/P_atm and splitting ratio sends fraction f through chamber, delivered fraction ≈ f × F_sat (idealised). Dial engineering implements the non-linear mapping because F_sat is large. [1]

Pre-use checks (with machine checklist)

  • Correct agent, filled (not overfilled), upright
  • Interlock works
  • No leaks on mount
  • Dial moves freely; zero when off
  • For desflurane: powered, warmed, no fault alarms
  • AAGBI machine check includes vaporiser seating and leaks [1]

SAQ scaffold

  1. Draw variable-bypass principle.
  2. SVP table and why desflurane special.
  3. Temperature compensation mechanism.
  4. Safety interlocks and keyed fillers.
  5. Draw-over vs plenum.
  6. Effect of tipping/overfilling (danger of liquid in bypass). [2]

Viva phrases

  • "What happens if you tip a full vaporiser?" → "Liquid agent may enter the bypass, delivering a potentially lethal concentration — take out of service and service protocol."
  • "Why not put desflurane in a sevo vaporiser?" → "SVP and boiling point make output dangerously high and uncontrolled; keyed fillers prevent it." [3]

Common traps

  • Filling sevo vaporiser with isoflurane.
  • Leaving vaporiser on during machine check flush without awareness.
  • Using desflurane vaporiser unplugged.
  • Forgetting interlock failure scenarios with two agents. [1]
Splitting ratio schematic
FigureBypass versus chamber flow: the dial is a calibrated splitter.
~157 mmHg
Sevo SVP 20 °C
~669 mmHg
Des SVP 20 °C
22.8 °C
Des boiling point
Splitting ratio
Plenum idea
[1]

Variable-bypass plenum

  • Sevo/iso/halothane
  • Temp compensated
  • Agent-specific
  • Flow-resistant

Desflurane special

  • Heated ~39 °C
  • Pressurised
  • Electronic
  • Needs power

Draw-over

  • Low resistance
  • Field use
  • Less perfect compensation
  • Manual skill

Safety systems

  • Keyed filler
  • Interlock
  • Level window
  • Anti-wrong-agent
[1]

Definition

Desflurane boils at room temperature — that single physical fact forces an entirely different vaporiser architecture. [1]

Clinical pearl

If the end-tidal agent is zero and the vaporiser is "on," think empty sump, wrong mount leak, or (for des) power/warm-up fault before blaming the monitor alone. [1]

Integrated exam drill sheet

Sixty-second version

Say the definition, the critical number or sequence, the main clinical use, and the top red flag. Stop. If you cannot do this without notes, the topic is not yet learnable.

Three-minute version

Add mechanism, a comparison table spoken aloud, one special population, and one crisis stem with first actions. This is the standard viva unit.

Ten-minute mastery version

Add equipment detail or procedural steps, evidence limits, second-line options, and a teach-the-junior summary. This is Final long-case depth.

Written SAQ timing

For a 10-minute SAQ, spend one minute planning headings, seven minutes writing, two minutes checking hard stops and units. Headings should mirror examiner dimensions: definition, mechanism or anatomy, clinical application, complications, special situations.

Common mark-losing behaviours

  • Lists without mechanisms
  • Mechanisms without clinical action
  • Doses without route or monitoring
  • Landmarks without injury consequences
  • Device talk without re-enable or backup plans
  • Absolute claims where practice is protocol-dependent

Positive mark-gaining behaviours

  • Numbers with units and approximate ranges
  • Explicit assumptions for equations
  • Side-by-side comparisons
  • Named hard contraindications
  • Monitoring endpoints
  • Clear escalation

Cross-specialty board alignment

ANZCA Primary and Final, FRCA Primary and Final, ABA, EDAIC and FCAI all test these leaves repeatedly because they are portable across subspecialties. A candidate who owns flow physics, electrical safety, neck and neuraxial anatomy, vaporiser principles and core adjunct pharmacology can survive stems in ICU transfer, obstetric haemorrhage, thoracic lists and outpatient dental anaesthesia alike.

Personal rehearsal script

Read the AnswerCard twice. Cover it and rewrite it from memory. Speak the red flags. Draw one table from memory. Answer one hostile interruption. Then move on. Spaced repetition beats marathon re-reading.

Safety culture close

Every technical topic ends in patient safety: do not expand closed gas spaces, do not dilate arteries, do not leave ICD therapies off, do not apply Poiseuille in turbulence, do not ignore conus level, do not tip a full vaporiser back into service without protocol, and do not stack serotonergic weak opioids casually. Knowledge is only exam-pass when it prevents harm.

Topic-specific mastery addendum

Layered recall sequence

  1. Recite the AnswerCard from memory.
  2. Draw the key table (layers, nerves, or vaporiser types) from blank paper.
  3. Speak two viva stems with full answers.
  4. List every red flag without looking.
  5. Teach the complication map as a chain: error → injured structure → clinical syndrome → immediate management.

Procedural narration standard

Narrate as if a consultant is watching: position, asepsis or machine check, landmark or ultrasound view, needle or dial action, endpoint, confirmation test, contingency if endpoint missing, and documentation. This narration style scores in OSCE-like and viva settings because it proves usable competence rather than passive recognition.

Numbers and relations to keep hot

Keep a personal card of the five hottest facts for this leaf and revisit them daily for a week. For neuraxial spaces that means conus level, Tuffier estimate, midline tissue sequence, epidural contents and neonatal difference. For cranial nerves that means V/IX/X airway map, SLN versus RLN, oculocardiac pathway and Horner triad. For vaporisers that means splitting ratio, SVP contrast for desflurane, temperature compensation idea, keyed fillers/interlocks and tip-over danger.

Error museum

Build an "error museum" of classic failures: spinal at L1/2 in adults; dilating an artery; leaving vaporiser tipped in service; missing bilateral RLN injury risk; assuming magnet behaviour is universal; ignoring delayed respiratory depression after neuraxial morphine. Each exhibit should have the false belief, the correct belief, and the protective habit.

Link-forward reading

After mastering this leaf, deliberately link to the next clinical topic in the same sitting: neuraxial anatomy to CSE and caudal; cranial nerves to awake FOI and eye surgery reflexes; vaporisers to volatile agent pharmacology and circle-system low flow. Linked encoding is more durable than isolated topics.

Worked vaporiser stems

Stem — end-tidal agent zero with dial on. Check empty sump, mount leak, wrong interlock position, or for desflurane power and warm-up faults before blaming the gas analyser.[1]

Stem — why desflurane needs a special vaporiser. Boiling point about 22.8 °C and SVP near atmospheric pressure mean a standard variable-bypass design cannot control output safely; heated pressurised injectors are required. [2]

Stem — tipped vaporiser. Liquid may enter the bypass and deliver a potentially lethal concentration. Remove from service and follow service protocol; do not “run it empty and hope.” [3]

Stem — altitude question. Clinical potency tracks partial pressure. Variable-bypass behaviour with altitude is often taught as roughly preserving partial pressure output more than volume percent; state the principle and avoid over-precise folklore. [1]

Stem — machine check. Confirm seating, interlock, filler integrity, level, and that only one vaporiser can open. AAGBI-style checks include vaporiser leaks and security on the back bar.[1]

Splitting ratio language: FGF divides into bypass and chamber streams; chamber gas is saturated at SVP; streams rejoin to the dialled concentration. Temperature compensation prevents cooling-related fall in output as vaporisation steals heat. [2]

Red flags

Red flag

Variable-bypass: splitting ratio + saturated chamber + temperature compensation. [1]

Red flag

Desflurane: heated, pressurised vaporiser — SVP near atmospheric; BP 22.8 °C. [1]

Red flag

Agent-specific keyed fillers; interlock one at a time. [1]

Red flag

Tipped/overfilled vaporiser can deliver liquid agent — lethal risk. [1]

References

  1. [1]AAGBI, Hartle A, Anderson E. Checking anaesthetic equipment 2012: association of anaesthetists of Great Britain and Ireland Anaesthesia, 2012.PMID 22563957
  2. [2]Tiviraj S, et al. Development of An Assessment Test for An Anesthetic Machine J Med Assoc Thai, 2016.PMID 27501622
  3. [3]Parthasarathy S, Ravishankar M. The closed circuit and the low flow systems Indian J Anaesth, 2013.PMID 24249885